Default THERMOTRON 8800 CONTROLLER

Anyone have tranducer problems with their 8800 controller? This chamber has been nothing but problems and they are all related to the controller. First it would run for about 10 days without a problem, then it would lose its mind and the process variable would be jumping all over the place, since its a new chamber thermotron replaced the main board. Now its a couple of months later and the transducer board is telling the CM that the discharge pressure of both systems is 600psi and both suction pressures are 200psi as well as we are out of water in the humidity generator. All of this information comes from the transducer board. Anyone have a quick fix for this? I will exchange your fix for the factory configuration code.
Thanks
Bob
Reply With Quote
  #2  
Old 05-05-2009
Devendra Devendra is offline
Junior Member
 
Join Date: Mar 2009
Posts: 10
Devendra is on a distinguished road
Default tHERMOTRON 8800 CONTROLLER

Transducer board may be sending incorrect signal or data in controller is corrupted. Usually transducer board send 4-20ma or 0-1V signal for system monitor. Check the supply signal and compare with the pressure range. Contact manufacturer for detail troubleshooting.
Reply With Quote
  #3  
Old 05-28-2009
bob9621 bob9621 is offline
Junior Member
 
Join Date: Apr 2009
Posts: 2
bob9621 is an unknown quantity at this point
Default thermotron8800 controller

We have had this thermotron since september 2008, we purchased it new from t-tron. Heres a list of what has had to be replaced so far
Cascade condenser
circulator fan motor
condenser fan motor
8800 main board
transducer board
developed a leak in the high stage due to factory installation of discharge line resting on compressor pan edge
discharge pressure solenoid valve
EPR valve for humidity cooling
low side discharge pressure regulator valve

Each of these repairs were done by t-tron service engineers under warranty so it only cost us down time and the cost of shifting our test to a lab with a chamber that we could rent. If I was to comment on the quality at thermotron I would have to give them a D. Only thing saving them would be that they did fix the problems at no charge to us.
Reply With Quote
Reply

Tags
8800 controller, thermotron, transducer board

Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump


All times are GMT -5. The time now is 11:11 PM.


Template Provided by Windows 7 Forums
Powered by vBulletin® Version 3.8.1
Copyright ©2000 - 2010, Jelsoft Enterprises Ltd.
Ad Management plugin by RedTyger